phil page Posted May 9, 2013 Posted May 9, 2013 Any advice about getting these old lamps through pat test.Passed electrically, reason for failure given lack of cord grip.I think the internal wiring should be renewed ...can I just do this and run it out to the existing flex using a ceramic terminal connector, or solder?I think its actually pretty well gripped inside although not obvious from the outside .... any ideas for a quick fix that will look acceptable. cheerPhil
jonathanhill Posted May 9, 2013 Posted May 9, 2013 I thought that the Patt. 23 had an internal cord grip. If you have to re-wire, then silicone cable or a fibreglass sheath must be used. Ceramic terminal blocks should be used if a cable has to be joined. Solder should not used within the body of a lantern. Who has failed these lanterns? Is there scope for a retest in their current condition?
Don Allen Posted May 10, 2013 Posted May 10, 2013 Item 12 on the exploded diagram of the Patt 123 is cable clamp. From Strand Archive Patt 123 specification "A strain relief gland shall secure 30-in. (0.75m) external length of 3-conductor heat-resisting sheathed cable with earth/ground conductor"
sguy42 Posted May 10, 2013 Posted May 10, 2013 It depends on the age of the Patt 123. If they are from mid 1967 onwards then they should have an external cable grip, although designs changed slightly with time. Earlier units (which originally had 3 seperate leads for live, neutral and earth)had an internal clamp for live and neutral, which wouldn't be considered suitable today. It is possible to fit a modern cable grip to these units but a little metal work is required to drill out the hole and ensure that there is enough clearence for the focus mechanism.
ianknight Posted May 10, 2013 Posted May 10, 2013 It depends on the age of the Patt 123. If they are from mid 1967 onwards then they should have an external cable grip, although designs changed slightly with time. Earlier units (which originally had 3 seperate leads for live, neutral and earth)had an internal clamp for live and neutral, which wouldn't be considered suitable today. It is possible to fit a modern cable grip to these units but a little metal work is required to drill out the hole and ensure that there is enough clearence for the focus mechanism. The usual place you find after market glands fitted is at the rear of the bulbous undersection on either the left or the right side - take care to centre your bit when drilling or the locking nut for the gland won't fit :P (I only messed two up - honest!)
emsgeorge Posted May 10, 2013 Posted May 10, 2013 And don't forget about the asbestos on some of the very early cables ! !.
paulears Posted May 10, 2013 Posted May 10, 2013 I find it mildly amusing that something reliable enough to still be in service after this amount of time can be failed because of a poor cable clamping design - er, 40 years later, we decide it's no good!
ianknight Posted May 10, 2013 Posted May 10, 2013 I find it mildly amusing that something reliable enough to still be in service after this amount of time can be failed because of a poor cable clamping design - er, 40 years later, we decide it's no good! Even more amusing that folk want to keep a 40 year old instrument working rather than pension them off!
sleah Posted May 10, 2013 Posted May 10, 2013 I find it mildly amusing that something reliable enough to still be in service after this amount of time can be failed because of a poor cable clamping design - er, 40 years later, we decide it's no good! Even more amusing that folk want to keep a 40 year old instrument working rather than pension them off! If it ain't broke..... ;) Oh wait... :blink:
phil page Posted May 11, 2013 Author Posted May 11, 2013 thanks folks. I'm a little clearer ......... they are of different dates, I have a few with external cable grip, which we are still using.I'm afraid budget cuts don't allow me to retire any of these ..... as long as there safe. I doubt the technician looked inside them, just plugged them into his test unit, and is unaware that they have an inside grip. I feel the crumbly insulation is more of a concern than the cable grip.I could do a bit of work on them. Where would I get a suitable cable grip ...... searches are confusing these things seem to have various names. Phil
emsgeorge Posted May 11, 2013 Posted May 11, 2013 type the model number into ebay. You will see some of the prices they go for, being 'vintage'. You could buy a lot of replacements !
Don Allen Posted May 12, 2013 Posted May 12, 2013 Look for a metal cable gland like this Farnell UK the correct internal size to suit your cable.
Don Allen Posted May 13, 2013 Posted May 13, 2013 Look for a metal cable gland like this Farnell UK the correct internal size to suit your cable. Here is a more affordable range Altronics cable glands
phil page Posted May 13, 2013 Author Posted May 13, 2013 Yes .... thanks that looks like what I need Phil
ianknight Posted May 14, 2013 Posted May 14, 2013 type the model number into ebay. You will see some of the prices they go for, being 'vintage'. You could buy a lot of replacements ! I hear what you're saying but I prefer the large diameter Fresnel lens of the 123 against that on the Acclaim or Minuette - it seems to give a better field (but then that may be a wholly subjective judgement based on the fact I grew up with the things!).